Aladdin’s turbidity standards for ratio and non-ratio instruments are composed of suspended polymer microspheres. These turbidity standards remove the handling, stability and accuracy problems associated with traditional Formazin turbidity standards
Specifications
Specifications & Purity
Description:Non Ratio 10 NTU
Storage
Room temperature
Shipped In
Normal
Advanced Data
Certificates (CoA,COO,BSE/TSE and Analysis Chart)
C of A & Other Certificates(BSE/TSE, COO):
Analytical Chart:
Find and download the COA for your product by matching the lot number on the packaging.
Determine the necessary mass, volume, or concentration for preparing a solution.
Dilution Calculator
Determine the dilution needed to prepare a stock solution.
Reconstitution Calculator
Reviews
Customer Reviews
Application Protocols
General protocol for calibrating a nephelometric turbidimeter (illustrative; adapt to your SOP and the instrument manual):
Preparation
Verify instrument cleanliness and warm-up. Confirm the method (EPA 180.1 or ISO 7027) and the compatible unit scale (NTU/FNU). Inspect standards for expiration and integrity.
Mixing and temperature control
Equilibrate standards and cells to the measurement temperature (20–25 °C). Gently invert standards several times; avoid bubbles.
Blank and zero
If a blank is specified by your method, measure the blank (e.g., high-purity water) and set the instrument zero.
Calibration points
Measure standards from low to high turbidity. For each point, rinse the cell with a small aliquot of the standard, fill to the mark, cap, and wipe exterior. Insert with consistent orientation. Record the displayed value and temperature.
Curve fitting and verification
Apply the instrument’s calibration routine. Verify with an independent check standard not used for calibration. Results should fall within your acceptance criteria or within the standard’s stated uncertainty.
Documentation
Record lot numbers, assigned values ± uncertainties, expiration, and operator ID. Save calibration data per QA requirements.
Item-specific assigned values, uncertainty, and dilution instructions are not provided here; follow the CoA/Spec Sheet supplied with the product.
Biological Roles
Not typically applicable. Turbidity standards are analytical calibration materials and do not have inherent biological roles or metabolic functions.
Relevant context (general):
In microbiology and environmental science, turbidity measurements can correlate with particulate content, including microbial cells or suspended solids in water. However, the standards themselves are not biological reagents; they provide a reference scattering intensity for instrument calibration.
If a biocide is present in a standard (not specified here), it serves only to maintain product integrity by inhibiting microbial growth in the bottle. It does not imply any biological activity beyond preservation.
For biological assays requiring turbidity-derived estimates (e.g., McFarland equivalents for bacterial suspensions), use the assay-specific standards/protocols. This product’s suitability for such use is not specified; refer to the CoA/Spec Sheet and method requirements.
Buffer Applications
Not applicable. Turbidity standards are not buffering agents and are not used to prepare pH buffer systems.
Practical note (general):
If dilution is permitted by the CoA, use high-purity, low-particulate water. Avoid buffers, salts, or surfactants that can alter light scattering. Maintain neutral conditions unless the manufacturer specifies otherwise.
Green Alternatives
Context: Turbidity standards are used for instrument calibration rather than chemical transformations, but there are still sustainability and safety considerations.
Greener choices and practices (general/literature):
Ready-to-use vs. in-lab preparation: Purchasing stable, ready-to-use turbidity standards can reduce hazardous waste and analyst exposure compared with preparing formazin in-house from hydrazine sulfate and hexamethylenetetramine (which involve toxic or sensitizing reagents). Tradeoff: commercial products have packaging waste; select appropriate unit sizes to minimize leftover volumes.
Polymer-bead standards: Where method-acceptable, polymer microsphere-based standards can avoid aldehyde/hydrazine chemistry associated with formazin. Tradeoff: optical response may differ by wavelength; ensure compatibility with EPA 180.1 or ISO 7027 as required.
Shelf-life optimization: Choose package sizes aligned with usage to minimize discard of expired standards. Implement first-expire, first-out (FEFO) inventory rotation.
Water and cleaning: Prefer non-abrasive, low-foam cell cleaning procedures and ultrapure water rinses to reduce detergent use. Reuse cells instead of disposables where validated.
Transportation footprint: Consolidate orders and choose ambient-stable standards to avoid cold-chain shipping when product specifications allow (this item ships "Normal" and stores at room temperature per Product Data).
Item-specific formulation (e.g., formazin vs. polymeric) and method compatibility are not specified for this product; confirm on the CoA/SDS before implementing any green substitution strategy.
Pharmaceutical Uses
No therapeutic or clinical application. Turbidity standards are used solely for analytical calibration and verification.
Industry context (general/non-clinical):
Utilities monitoring: Used to calibrate instruments that monitor turbidity in purified water, water for injection (WFI), and process water in GMP settings, where turbidity is a quality attribute for filtration performance and cleanliness (per site-specific specifications and pharmacopeial guidance). Compatibility with specific compendial methods is method- and instrument-dependent.
Manufacturing QC: Employed to verify instrument linearity and sensitivity during equipment qualification (IQ/OQ/PQ) and routine calibration. Documentation should capture lot number, assigned value ± uncertainty, expiration, and environmental conditions.
No excipient role: These standards are not formulation excipients.
Item-specific regulatory status (e.g., traceability to compendial or NIST references) is not provided here; consult the CoA/Spec Sheet and your quality unit before use in regulated workflows.
Physical Properties
Item-specific properties from Product Data:
Appearance: Not specified for this item; refer to CoA/Spec Sheet.
Density, refractive index, pH, preservative content: Not specified for this item; refer to CoA/Spec Sheet.
General properties of turbidity standards (literature/general; not item-specific):
Physical state: Typically aqueous solutions/suspensions packaged in sealed vials or bottles. The matrix is deionized water; the turbidity arises from suspended polymeric microspheres or from a formazin colloid.
Stability: Commercial standards are designed to be stable for months to years when unopened and stored at recommended conditions. Once opened, stability may be limited by contamination or evaporation; always rely on the product’s stated expiration.
Turbidity values: Assigned values are reported in NTU/FNU/EBC, depending on the applicable method. Standards are commonly supplied at multiple levels (e.g., low single-digit up to several hundred NTU) for multi-point calibration checks. The exact values for this item are not specified; see CoA.
Temperature effects: Turbidity readings are temperature dependent (scattering changes with solvent density/viscosity). Calibration and measurement should be performed at a controlled temperature per method (often 20–25 °C) for best reproducibility (general guidance).
Chemical compatibility: Aqueous standards can be affected by surfactants or solvents; avoid contact with organic residues in cuvettes which can alter scattering behavior (general lab practice).
Note: Do not treat the assigned turbidity as a physicochemical constant of a molecule; it is a metrological quantity with uncertainty and traceability captured on the CoA.
Quality and Grades
Item-specific quality information from Product Data:
Grade/Purity: Not specified for this item; refer to CoA/Spec Sheet.
How quality is typically defined for turbidity standards (general/literature):
Assigned value and uncertainty: The primary quality attributes are the nominal turbidity value(s), the measured value(s) with uncertainty, and the traceability chain (e.g., to primary formazin reference or to recognized reference materials/procedures). This is documented in the CoA.
Method compatibility: Standards may be designated for EPA 180.1 (NTU, white light), ISO 7027 (FNU, infrared), or instrument-specific optics. Using a standard on the wrong optical system can introduce bias.
Stability and shelf life: CoA should state expiration, storage temperature, and whether the container is single-use or multi-use. Unopened stability differs from in-use stability.
Homogeneity: For multi-bottle sets, between-bottle homogeneity is critical; reputable suppliers validate this by measurement and include it in uncertainty budgets.
Matrix and preservative: Some standards are preservative-free; others contain low-level biocides. Presence/absence influences in-use shelf life and handling restrictions. For this item, composition is not specified.
Verify the measurement scale and traceability on the CoA before use.
Record lot number, assigned value ± uncertainty, temperature, and instrument ID during calibration/verification.
Replace standards when expired or if contamination/cloudiness/precipitation inconsistent with the nominal value is observed.
Reaction and Applications
This product is not a reagent for synthetic reactions; it is an analytical calibration/verification standard. Relevant applications include:
Instrument calibration (general): Establish multi-point calibration curves for turbidimeters or nephelometers. Use at least three levels spanning the measurement range (e.g., low, mid, high). Follow your method (EPA 180.1, ISO 7027, Standard Methods 2130 B) for frequency and acceptance criteria.
Daily/shift checks (QA/QC): Verify instrument performance using a mid-level check standard at defined intervals. Track results on control charts to detect drift or fouling of optics.
Method validation/verification: Use standards to evaluate linearity, precision, and accuracy of turbidity measurements across the intended range. Confirm carryover and sample cell effects.
Training and proficiency: Provide consistent materials for analyst training and proficiency testing in water, beverage, pharmaceutical utilities, and environmental laboratories.
Sample handling studies: Assess the effect of temperature, aeration, and mixing on turbidity readings to optimize SOPs for field and bench measurements.
Practical tips (general):
Mix by gentle inversion; avoid shaking that introduces microbubbles. Wait briefly for bubbles to dissipate before measurement.
Rinse cells with a small aliquot of the standard; wipe the outside with a lint-free cloth to remove fingerprints/water droplets.
Record the lot’s assigned value, uncertainty, expiration, and temperature. Apply stray-light correction or instrument-specific factors only as prescribed by your method.
Item-specific assigned values and optical method compatibility are not provided here; consult the product’s CoA/Spec Sheet.
Reaction Conditions
Not applicable to chemical reactions. However, for the analytical use of turbidity standards, the following measurement conditions are critical (general guidance):
Temperature: Calibrate and measure at a controlled temperature, typically 20–25 °C, as scattering intensity can vary with temperature and viscosity.
Mixing: Gently invert several times to achieve homogeneity. Avoid shaking that introduces microbubbles. Allow entrained bubbles to dissipate before measurement.
Optical system compatibility: Match the standard to your instrument’s method (e.g., EPA 180.1 NTU vs ISO 7027 FNU). Using mismatched standards can bias results.
Cell handling: Use clean, unscratched cells. Rinse with a small aliquot of the standard and wipe the exterior with lint-free tissue.
Acceptance criteria: Establish calibration/verification acceptance windows per your method/SOP (e.g., within stated uncertainty or a percentage of the assigned value). Document results and investigate out-of-tolerance findings (dirty optics, lamp aging, contamination).
Item-specific instructions (dilution factors, equilibration times, assigned values) are not provided in the Product Data; refer to the product CoA/Spec Sheet and instrument manuals.
Safety and Handling
Item-specific safety from Product Data:
GHS Classification: Not specified for this item; refer to SDS.
Signal Word: Not specified for this item; refer to SDS.
H-Statements / Pictograms: Not specified for this item; refer to SDS.
General safety guidance for turbidity standards (consult the product SDS for authoritative information):
Composition: Typically water-based; may contain dispersants or preservatives (e.g., low levels of biocide) depending on formulation. As the exact formulation is not provided here, treat as a laboratory chemical and avoid ingestion/inhalation/skin or eye contact.
PPE: Wear safety glasses, lab coat, and appropriate chemical-resistant gloves. Use good laboratory hygiene (wash hands after handling, avoid pipetting by mouth).
Handling: Mix gently by slow inversion to resuspend any settled material; do not vortex aggressively to avoid microbubble formation which can alter turbidity readings. Prevent microbial contamination by minimizing open time and avoiding contact between containers and instrument cells.
Incompatibilities: Avoid strong oxidizers, strong acids/bases, and organic solvents that can destabilize the suspension. Do not autoclave unless explicitly allowed by the manufacturer.
First aid (general): If on skin or in eyes, rinse with water for several minutes; remove contact lenses if present and easy to do. If inhaled (aerosols), move to fresh air. If ingested, rinse mouth and seek medical advice. Always follow the SDS.
Spills and disposal: Wipe small spills with absorbent and water. Dispose of waste according to institutional and local regulations; do not discharge into drains without authorization.
Storage condition from Product Data: Room temperature. Shipments are "Normal" per Product Data. Protect from freezing unless the CoA/SDS indicates otherwise.
Solvent Selection
Applicability: Turbidity standards are typically aqueous-based. Classical "solvent selection" for reaction chemistry does not apply.
Practical guidance for use (general):
Matrix: Most standards are prepared in high-purity water. If dilution is required (check the CoA), use freshly produced, low-particle, low-TOC deionized water (e.g., 0.2 µm filtered) to avoid introducing scatterers.
Miscibility: Aqueous standards should not be mixed with organic solvents or surfactants, which can change scattering characteristics and bias readings.
Temperature control: Conduct calibration and verification at a consistent temperature (commonly 20–25 °C) to minimize temperature-dependent scattering differences.
Cuvette/cell compatibility: Use clean, scratch-free glass or polymer cells compatible with the instrument optics (EPA 180.1 vs ISO 7027). Residual films (detergents, solvents) can alter readings; rinse thoroughly with ultrapure water.
When to choose alternative matrices (literature/general):
For infrared (ISO 7027) instruments, use standards designated for FNU; for tungsten-lamp nephelometers (EPA 180.1), choose NTU-designated standards. Cross-use can produce systematic biases due to wavelength-dependent scattering.
Item-specific solvent/matrix details: Not specified for this item; refer to CoA/Spec Sheet.
Storage and Reconstitution
Item-specific storage/shipping from Product Data:
Storage Conditions: Room temperature.
Shipped In: Normal.
Additional item-specific details (stabilizers, freeze–thaw tolerance, in-use shelf life): Not specified for this item; refer to CoA/Spec Sheet and SDS.
General handling guidance for turbidity standards (not item-specific):
Storage: Keep tightly closed when not in use. Protect from freezing and excessive heat or direct sunlight unless the label explicitly allows it. Store upright to minimize cap leakage and evaporation.
Mixing before use: Gently invert to resuspend any settled material. Do not vortex; avoid creating bubbles that affect readings.
In-use practices: Minimize headspace exposure and avoid touching the bottle rim with cells or pipette tips to reduce contamination risk. If the standard becomes visibly contaminated or results drift, discontinue use.
Reconstitution/dilution: Many standards are ready-to-use. If the CoA specifies dilution (e.g., to prepare intermediate levels), use freshly produced, particle-free, high-purity water. Mix by gentle inversion and allow bubbles to dissipate.
Label control: Track open date and assign an in-use expiration per your QA policy or the manufacturer’s recommendation.
Always prioritize the instructions, assigned values, uncertainty, and expiration stated on the product’s CoA/Spec Sheet and SDS.
Structure and Identity
Brief overview: This product consists of turbidity calibration/verification standards intended for nephelometric or turbidimetric measurements. These are typically aqueous suspensions/solutions with defined turbidity values rather than a single, well-defined small molecule.
Item-specific identifiers from Product Data:
SKU: T197196
Product Name: Turbidity Standards
CAS: Not specified for this item; refer to CoA/Spec Sheet.
Molecular Formula: Not specified for this item; refer to CoA/Spec Sheet.
Molecular Weight: Not specified for this item; refer to CoA/Spec Sheet.
InChIKey: Not applicable for a mixture/standard set.
SMILES: Not applicable for a mixture/standard set.
Structural description (general knowledge):
Turbidity standards are commonly either polymeric bead dispersions or formazin-based colloids prepared in water; both are colloidal systems, not discrete molecules.
The analytical identity is defined by the assigned turbidity value (e.g., in NTU, FNU, EBC) and its traceability to a primary reference procedure or standard, rather than by specific chemical structure.
Traceability/conformity (literature/general):
Turbidity scales may follow regulatory/standard methods such as EPA 180.1 (nephelometric NTU, tungsten lamp), ISO 7027 (formazin nephelometric units, infrared), or instrument-specific units. The applicable scale for this item is not specified; consult the CoA/Spec Sheet.
Note: Because this product is a calibration standard set, classical molecular identifiers (SMILES, InChI) do not apply. Identity should be verified from the certificate of analysis (assigned value, uncertainty, and traceability).
Synthetic Utility
Not applicable to chemical synthesis. Turbidity standards are calibration materials and do not participate as reagents, catalysts, or solvents in synthetic transformations.
Related laboratory utility (general):
Method development: When turbidity interferes with spectrophotometric assays, standards can be used to model scattering backgrounds and test correction algorithms.
Filtration studies: Assess filter performance and retention characteristics by measuring turbidity reduction across filter media under controlled conditions.
For reagents intended for chemical synthesis, please refer to other product categories in our catalog.
Target Specificity
Not applicable. This product is not a biochemical probe, antibody, or affinity reagent and has no target specificity parameters (e.g., antigen, epitope, clone, isotype).
Frequently Asked Questions
How should this product be stored?
Store at room temperature.
How is this product shipped?
This product ships under standard ambient conditions. No temperature-controlled packaging is required.
What documentation is provided?
Available product documentation, including Certificates of Analysis (COA), Safety Data Sheets (SDS), and specification sheets, is shown in the product document area. Document availability and access follow the current site policy.
We use cookies to ensure the website functions properly and, where permitted, to improve your experience. You can manage your preferences at any time in Settings. Learn more in our Cookie Policy.
Shall we send you a message when we have discounts available?
Remind me later
Thank you! Please check your email inbox to confirm.
Products are supplied to verified businesses, institutions, and qualified professionals for research and development use only. Not for use in humans, animals, diagnosis, or therapy.